Change in GC (IRMS) chromatogram after maintenance

Discussions about GC and other "gas phase" separation techniques.

18 posts Page 2 of 2

Hi Alice

When you did the earlier maintenance, and the recent column change, did you change the gold seal at the bottom of the inlet ? Tailing with a brand new column is very likely to be dirt somewhere in the inlet.

Also, when you installed the new column did you trim a few cm off the end after you threaded it though the ferrules ?

Peter
Peter Apps

Hi Peter

I was going to inspect the gold seal, but the reducing nut was on so tight I was unable to take the entire assembly out. I will try again though.

Yes, I did trim some segment off the column after it was threaded through. I could see some of the graphite from the ferrule go into the column. Thanks.

Alice

Hi Alice

If the nut is on that tight it has probably never been removed, and you have 4 years of accumulated gunk sitting inthe bottom of the inlet - that would be enough to account for the problems.

Peter
Peter Apps
18 posts Page 2 of 2

Who is online

In total there are 40 users online :: 2 registered, 0 hidden and 38 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahrefs [Bot], Google [Bot] and 38 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry