- Posts: 345
- Joined: Sun Dec 06, 2009 6:16 pm
For UV the upper limit will be determined by the linearity of your particular detector. A general rule (from spectroscopy theory) is that maximum absorbance values below 1 AU (1000mAU) will still be in the linear region. Some vendors claim linearity above 1 AU but I have not tested this claim. Maybe someone else has. The low end will also be detector specific and is determined by the S/N. A S/N of at least 20 at the low end of your curve and absorbance below 1AU should give you a linear curve. However, by definition/theory prediction %error will be highest at the low end of the curve if no weighting is used.In my case, I use following concentration range: From 0,039 to 10 µg/ml (9 calibration points). Maybe this is a too large range for only one linear calibration curve and then the best solution is to use a low and a high calibration curve, as I already mentioned.
Does someone has any clue about, how large a concentration range can be, in order to get excact results over the whole range (UV detection) with only one linear equation?
I have had good success with 1/x^2 weighting. When looking at fit errors you must focus on %error and NOT absolute error. With 1/x^2 weighting typically the %error across the entire calibration range is similar unless there is some non-linearity in your system. Non-linearity could arise from several different sources, e.g. interference at low end or detector saturation at high end.