The v/v% viscosity maximum for ACN/H2O is ~24% ACN. What is the intial %ACN in your mobile phase?
Advertisement
Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.
The v/v% viscosity maximum for ACN/H2O is ~24% ACN. What is the intial %ACN in your mobile phase?
So this is a HILIC method. What column are you using?
With HILIC you'll need to reduce the water and remove the salts in your sample. The sample solvent strength (% ACN) and inj volume tolerated depend on the column dimensions (50x4.6mm ID?).
Can you afford to dilute the sample 4:1 with ACN (i.e. to 80% ACN)? If so, filter the diluted sample. Hopefully you wont lose any analyte thru co-precipitation.
Also, you can start/end your gradient at 85% ACN instead of 95% to speed up re-equlibration and to reduce run time and precipitation problems.
You're welcome! I'm happy to help a fellow chromatographer.The column is Agilent Luna NH2 100A, 5 um, 4.6x250mm.
Dr. Neue, that's an interesting point to consider. Luna Amino is not standard amino propyl chemistry and does not share their bleed charateristics.Standard amino columns bleed like crazy during the initial runs, and this can be the cause of clogged outlet frits.