3 µl of a 0.2 mg/ml solution on a 4.6 x 100 mm column?

I can only repeat myself. Note also, if this thing dissociates there could also be a severe pH mismatch in this method.

Hi Candy

If you can, remove or reduce the acetic acid in the sample solvent - it is probably causing a pH dispruption at the head of the column.

Peter
Peter Apps
Yes there was a huge shift from 23-3min, I have changed injection volume up to 3 microlitres, there was no change.
Are you suggesting that I use pure water instead?
Did the retention decrease from 23min at 80% MeOH to 3min at 60% MeOH?

Why is the acetic added to the sample diluent (MeCN, 0.5% acetic acid)

This molecule appears to be neutral so buffer additives should not have much of an effect.
A. Carl Sanchez
Yes there was a huge shift from 23-3min, I have changed injection volume up to 3 microlitres, there was no change.
Are you suggesting that I use pure water instead?
Did the retention decrease from 23min at 80% MeOH to 3min at 60% MeOH?

Why is the acetic added to the sample diluent (MeCN, 0.5% acetic acid)

This molecule appears to be neutral so buffer additives should not have much of an effect.
Yes the RT decreased from 23min at 80%MeOH to 3.9min at 60%meOH at both 1ml/min.
I add 0.5%acetic acid to stabilize the rapamycin and right now i'm trying it without the acetic acid
Yes the RT decreased from 23min at 80%MeOH to 3.9min at 60%meOH at both 1ml/min.
I add 0.5%acetic acid to stabilize the rapamycin and right now i'm trying it without the acetic acid
So the compound is weakly acidic (keto/enol tautomers?) and the retention behavior is very unusual. This retention behavior is more like HILIC than reversed phase which I would not expect for such a hydrophobic compound.

I suggest you pre-mix the following mobile phases by hand and use only one channel on the LC to verify your system

85/15 MeOH/H2O
80/20 MeOH/H2O
75/25 MeOH/H2O

Please let me know what retention times and peak areas you observe. What detector are you using?
A. Carl Sanchez

Also, dissolve the sample in 80/20 MeOH/H2O for this test
A. Carl Sanchez

Hi Candy,

Any progress?
A. Carl Sanchez
23 posts Page 2 of 2

Who is online

In total there are 35 users online :: 2 registered, 0 hidden and 33 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Amazon [Bot], Semrush [Bot] and 33 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry