Advertisement

LC of cyanoacetamide

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

19 posts Page 2 of 2

Uve Neue,
you are right, UV 190 is to low for 50:50. PH =5.5
Unmqvar
I do not understand the abbreviations MB and BTW.

MB is a typo mistake for MP= mobile phase
BTW is "by the way"

If you want to see your analytes in UV, you either need to use a phosphate buffer or no buffer. Sorry for insisting that HILIC on silica might be the better solution.

Uwe Neue,

your advice is greatly appreciated. I`ll try and report.