What is your mobile phase?
-
- Posts: 426
- Joined: Mon May 19, 2008 10:09 am
Maybe it is coincidence, but if you take a look at about 200 nm, you see a dip in the spectrum at the beginning of the peak, which slowly disappears, and back show up at the end of the peak.
This could be a small impurity with a spectrum that only has absorbance in the region < 205nm.
If you inject less volume, the underlaying peak can be too small to be detected.
Just some thoughts....
Ace
This could be a small impurity with a spectrum that only has absorbance in the region < 205nm.
If you inject less volume, the underlaying peak can be too small to be detected.
Just some thoughts....
Ace
-
- Posts: 181
- Joined: Tue Oct 28, 2008 9:54 pm
Mobile phase is 85% [50mM phosphate buffer, 5mM octane sulfate at pH 2] 12% MeOH 3% ACN
MestizoJoe
Analytical Chemist and Adventurer
Venture Industries
Spider-Skull Island
Analytical Chemist and Adventurer
Venture Industries
Spider-Skull Island
-
- Posts: 2916
- Joined: Mon Aug 30, 2004 10:19 pm
At the very low wavelength, you may be seeing the solvation of your analyte with methanol.
-
- Posts: 2846
- Joined: Mon Aug 30, 2004 7:17 am
Uwe, do you mean "interference from MeOH"?
It looks clearly as if a "zero" of baseline was never done here, so the specs of mobile phase components interfere.
It looks clearly as if a "zero" of baseline was never done here, so the specs of mobile phase components interfere.
-
- Posts: 239
- Joined: Tue Nov 09, 2004 3:06 pm
It looks very much like unappropriate "baseline correction". The dip at 200 nm could simply be "calcultated" because to much solvent signal is subtracted.
Try to refine the baseline correction. Or don't use it and compared spectra at the beginning and end of the peak. Or use a spectra range starting from 210nm (below everything is highly overlaid bei solvent absortion and therefore doesn't contain much information anyway).
Alex
Try to refine the baseline correction. Or don't use it and compared spectra at the beginning and end of the peak. Or use a spectra range starting from 210nm (below everything is highly overlaid bei solvent absortion and therefore doesn't contain much information anyway).
Alex
-
- Posts: 61
- Joined: Sat Nov 15, 2008 10:29 pm
I don't have a lot of experience in PDA peak purity measurements...actually, I have none...but with respect to the variation to which you refer: the same spectra are obtained on each "side" of the peak maximum at 16.408.So, I took the UV spectrum from the PDA at several time points within the peak. What I found was variation in the spectrum. I guess this can be considered noise. Correct me if I am wrong.
Compare:
16.051 with 16.840
16.092 with 16.811
16.235 with 16.723
16.323 with 16.639
They look the same. Just an observation. Good luck.
I love deadlines. I like the whooshing sound they make as they fly by.
-
- Posts: 426
- Joined: Mon May 19, 2008 10:09 am
But if you compare the spectra at one side:I don't have a lot of experience in PDA peak purity measurements...actually, I have none...but with respect to the variation to which you refer: the same spectra are obtained on each "side" of the peak maximum at 16.408.So, I took the UV spectrum from the PDA at several time points within the peak. What I found was variation in the spectrum. I guess this can be considered noise. Correct me if I am wrong.
Compare:
16.051 with 16.840
16.092 with 16.811
16.235 with 16.723
16.323 with 16.639
They look the same. Just an observation. Good luck.
16.051 with 16.092 with 16.235 with...
Then you will find variation, which made me think that this is no random variation...
Ace
-
- Posts: 176
- Joined: Tue Aug 14, 2007 3:08 pm
MestizoJoe
Sir how could u insert images of PDA spectra
Sir how could u insert images of PDA spectra
-
- Posts: 65
- Joined: Wed Apr 11, 2012 1:48 pm
How could u have a small impurity that is much wider than the main compound?Maybe it is coincidence, but if you take a look at about 200 nm, you see a dip in the spectrum at the beginning of the peak, which slowly disappears, and back show up at the end of the peak.
This could be a small impurity with a spectrum that only has absorbance in the region < 205nm.
If you inject less volume, the underlaying peak can be too small to be detected.
Just some thoughts....
Ace
(You can see the same absorbance trace at 16 min and 16.9 min for that impurity?)
I think it cannot be one impurity.
-
- Posts: 65
- Joined: Wed Apr 11, 2012 1:48 pm
You said the only possible false positive is when the s/n is low.First of all, we have to realize the limitations of *any* PDA-based peak purity measurements. They will fail to detect an underlying peak (false negative) under four conditions:
1. The underlying peak has a very similar UV spectrum.
2. The underlying peak is very small.
3. The underlying peak has no UV absorbance over the wavelength range involved.
4. The underlying peak exactly coelutes with the peak of interest.
They will detect a non-existant underlying peak (false positive) only when the signal/noise ratio is low (very noisy baseline or very small peak of interest).
In this case (false positive?)Low-wavelength noise and/or saturation of the detector (as has been suggesting) is certainly a possibility. The other possibility is that you do, indeed have an underlying peak, which does not have a significant chromophore above 210 nm but *does* have an "end absorbance" below 210 nm (an aliphatic alcohol, for example).
What to do depends on your purpose. If its an assay (where you want to quantitate the main peak), then sitting at the absorbance maximum (presumably above 210 nm) would be fine. If you are looking for impurities, then there *is* a chance that you're missing something.
But my recent experiment of pure ref std shows that at higher conc, the purity fails (comparing purity factor and threshold) while at lower conc, it passes.
Both numbers are high 9999.xx something.
I am very confused.
-
- tom jupille
- Site Admin
-
- Posts: 4978
- Joined: Wed Aug 11, 2004 4:55 pm
I clarified it a bit in the third paragraph by including "saturation of the detector".
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
27 posts
Page 2 of 2
Who is online
In total there are 14 users online :: 1 registered, 0 hidden and 13 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Google [Bot] and 13 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Google [Bot] and 13 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science