Finally, how often is the instrument being tuned and is it being tuned with or without mass pattern adjustment?
You should also check the autosampler parameters. The viscosity delay defaults to 0.2 sec, which is shorter than I like. I prefer a 1 sec delay. All the syringe speed parameters should be set to High. Be sure and click on the Advanced button and check the injection port dwell time. The default is 0.3 sec, which can be an issue. I prefer 0, or if doing splitless injections at low concentration 1 to 2 sec.














 
																							