I went back and reviewed chromatography for the last few months of runs. There's some variability in the magnitude of the downward drift, but it's always there. To me, this rules one-time errors (eg, incorrect reagent selection) which would have a more transient effect. That said, purity/contamination of the KH2PO4 and/or H3PO4 is definitely still on my list, as in my experience, these reagents can be in use for quite some time.One more thought... if the buffer solution was pH adjusted with the wrong acid (e.g. Acetic), then the addition of the impure acid with higher absorbance may also result in the same downward drift as the ACN composition is increased. In the case of acetic acid, it is often of low purity and goes bad quickly resulting in an observable brown color.
One other area I've been considering is the column being used (Waters Acquity BEH C18). I don't have a lot of experience with these 'Ethylene bridged hybrid' particles. I'm certain that the chromatographic conditions are well within the operating range of these columns, but I'm not sure if there are any special conditioning/equilibration procedures that should be followed when using them. For example, is there any reason that these hybrid particles might require longer equilibration periods, or a different post-run washing procedure?