522 1,4-Dioxane Sensitivity

Discussions about GC and other "gas phase" separation techniques.

20 posts Page 2 of 2
What type of column are you using for analysis? Are you following method oven parameters? What activated carbon cartridges are you using? We are currently using Waters Sep-pak and have also tried the Phenomenex cartridges as well, but it has not given much difference as far as the coelution of the M+ 58 ion. We also employ a promochrome automated SPE system and have not experimented with a manual SPE manifold just yet. Is your SPE done manually?
We haven't had any luck resolving the M+ 58 peak either, although we can get down to 0.1 ug/L for 1,4-dioxane before the ion ratio integration falls outside of the method spec.

(Previously, this had been consistently <0.04 ug/L in our lab for the 100 mL extraction option before something in the supplies changed). All of our solvent blanks and unextracted blanks are still clean. A large bolus of the unknown contaminant comes off when first wetting the carbon cartridge, then it seems clean until the methanol is in the extraction and then more comes elutes off.
What type of column are you using for analysis? Are you following method oven parameters? What activated carbon cartridges are you using? We are currently using Waters Sep-pak and have also tried the Phenomenex cartridges as well, but it has not given much difference as far as the coelution of the M+ 58 ion. We also employ a promochrome automated SPE system and have not experimented with a manual SPE manifold just yet. Is your SPE done manually?
Waters Sep Pak AC-2
Restek Rxi-624 Sil MS 0.25 mm (using the oven profile in the method)
Helium at 1 mL/min (through Agilent big universal trap)
Dichloromethane - Honeywell (B&J)
Methanol - Honeywell (B&J) P&T
1 uL injection size

We reviewed multiple vendors for our solvents and these are the best we found through actual testing looking for the interfering peak.
I have recently tried and found success on getting the confirmatory ion down using a RTX-VMS 30x0.25mmIDx1.4um column :)
I have recently tried and found success on getting the confirmatory ion down using a RTX-VMS 30x0.25mmIDx1.4um column :)
By success, you mean that you can separate the M+ 58 impurity on the Rtx-VMS column? That would be great news.
20 posts Page 2 of 2

Who is online

In total there are 29 users online :: 1 registered, 0 hidden and 28 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Google [Bot] and 28 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry