injection 1: peak ok/injection 10: tailing and smaller area

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

19 posts Page 2 of 2

I do have a question for you...

Does the retention time of your last impurity remains the same or changes over time? Does the intensity/peak area change when you equilibrate more your column or not? What about the main and the impurities retention time? Do they remain the same or drift to higher retention times over time (which could explain the tailing).

- the retention time of the last impurity does not change over time
- the peak area is independent on the equilibrium time
- the retention times of all other compounds (main and other imp) remain constant

[/img]

I just visited the URL of the manufacturer of the ion-pairing reagent. They claim to act as supplier for:

- paint thinners
- paints / lacquers
- cleaning agents

Thus, a change of supplier might really be a good idea...

If changing the supplier solves the problem – as you describe it, it would be more art than science. Apropos paint industry :wink:

Best Regards
Learn Innovate and Share

Dancho Dikov
19 posts Page 2 of 2

Who is online

In total there is 1 user online :: 0 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 1 guest

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry