Separation of 2 peaks elutaing at approimately the same RT
Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.
52 posts
Page 2 of 4
trying= varying pH, temperture and organic compostion.............this was mostly performed on the other columns as well
-
- Posts: 2916
- Joined: Mon Aug 30, 2004 10:19 pm
How about SymmetryShield?
-
- Posts: 212
- Joined: Fri Dec 05, 2008 11:03 am
yes........but the problem is the pencilloic acid has 2 peaks, one broad small peak at approx 8mins and the other at 12mins( narrow peak..good peak shape)......so they kinda stretch out over a 5 minute retention. the sodium benzoate is in and around this retention time...so if you separate it from one peak it usually runs into the other!!
-
- Posts: 212
- Joined: Fri Dec 05, 2008 11:03 am
we have no SymmetryShield in house...thanks Uwe for sending those papers on solvent and stationary phase selection
-
- grzesiek
that's kinda interesting 
is this the acid?
http://en.wikipedia.org/wiki/Penicillic_acid
is it included in pharmacopoeia as impurity? does the Ph states that there shoul be two peaks?
Anyone knows if existence of two forms is pH dependent?
is this the acid?
http://en.wikipedia.org/wiki/Penicillic_acid
is it included in pharmacopoeia as impurity? does the Ph states that there shoul be two peaks?
Anyone knows if existence of two forms is pH dependent?
-
- Posts: 212
- Joined: Fri Dec 05, 2008 11:03 am
no...its Penicilloic acid of phenoxymethylpenicillin...and yes is an EP impurity. the reason it has 2 peaks ..i think due to its isomer number 1,2 and 3 penicilloic acid.
-
- Posts: 5433
- Joined: Thu Oct 13, 2005 2:29 pm
What is the injection solvent ?
Peter
Peter
Peter Apps
-
- Posts: 212
- Joined: Fri Dec 05, 2008 11:03 am
Phenoxymethylpenicillin Potassium Imp. E (EP): (4S)-2-[Carboxy[(phenoxyacetyl)amino]methyl]-5,5-dimethylthiazolidine-4-carboxylic Acid (Penicilloic Acids of Phenoxymethylpenicillin)
Hope this helps
Hope this helps
-
- Posts: 212
- Joined: Fri Dec 05, 2008 11:03 am
injection solvent/Diluent= 6.81g KH2PO4 in 1000ml DH20 with adjusted pH with 0.2m NaOH to pH6.5
-
- Posts: 1408
- Joined: Thu Jul 28, 2005 2:08 pm
Now, why would the pH of the sample solvent be 6.5 whilst the mobile phase’ pH is intended to be 3.5?
I write “intendedâ€
I write “intendedâ€
Learn Innovate and Share
Dancho Dikov
Dancho Dikov
-
- Posts: 212
- Joined: Fri Dec 05, 2008 11:03 am
Sorry...i spent the last 30mins trying to transfer the structure to the post..i couldnt do it. I can give you link
http://www.lgcstandards.com/media/559600807.pdf
you will have to download the catalogue...go to page 312 to see its structure.
About the difference of pHs of moblie phase and diluent...well thats the way they have it in the EP!...i thought it was a bit strange as well!
http://www.lgcstandards.com/media/559600807.pdf
you will have to download the catalogue...go to page 312 to see its structure.
About the difference of pHs of moblie phase and diluent...well thats the way they have it in the EP!...i thought it was a bit strange as well!
-
- Posts: 212
- Joined: Fri Dec 05, 2008 11:03 am
injection volume 20uL
52 posts
Page 2 of 4
Who is online
In total there are 35 users online :: 1 registered, 0 hidden and 34 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Google [Bot] and 34 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Google [Bot] and 34 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science