Posted: Fri Oct 09, 2009 1:15 pm
A silly question, do you premix the whole solvent. Or have you your system mixing the aqeous phase with acetonitrile?
Chromatography Forum is a public discussion group where you can post questions, news, or messages of interest to chromatographers everywhere, but you must be registered to participate (registration is FREE). If you are a registered user, please log in.
http://www.chromforum.org/
Dear Danko,Hi Ace,
How many consecutive injections is your observation based on? And is this 0.4 min a consistent retention time decreasing or does the retention time go the other way (increasing) as well?
Finally, when you removed the sodium sulfate from the mobile phase, did you lose the retention of polymyxine B completely (i.e. eluted with the void volume) or did it just get shorter and kept decreasing with every consecutive injection?
Best Regards
Hi Dirk,Dear Ace,
I looked up the EP and the state that for the method should be used a base deactivated and end capped C18 column. On the other hand you mentioned that Supelcosil ABZ was recommended which is a polar embedded C18. I believe the BEH Shield is also a polar embedded phase.
Do you have any other C18 phases in the lab that you could try? I doubt that the columns you are testing having a problem, but the mobile phase together with a polar embedded group might create some additional interactions that are harder to control. I believe my suggestion is aiming in the same direction that Uwe Neue gave.
Dirk
Lot of questions, I will answer them one by one:I am not only puzzled, but also confused. Did the retention time always change up and down, or sometimes only down? Did the rt variation (up or down or both) occur with all the columns and mobile phases tried? The corticoids were always present and never showed this rt variance?
Also, what is this 50mmol phosphate buffer + sodium phosphate?
Your analyte is a mixture? Did it give the same pattern of peaks under all conditions? Did the pattern change with rt changes under the same conditions?
Are you shure that the corticoids didn´t have a short enough rt to obliterate rt changes for them?
I don´t know anymore whether these were isocratic runs, if they were, why dont you premix the mobile phase and see what happens?
Another side remark: It is very puzzling to see many times, as here also, that people ask for help, but dont bother to answer all the question of people who are willing to assist.
One more question:
Do you have a possibly strong incompatibility between the pH/ion strength of sample solvent and mobile phase?
I wonder how it’s been confirmed. By omitting the sodium sulfate and observing considerable retention loss?I'd be less keen to continue to play with buffer strength, once you've confirmed that it's not fixing the problem.