Try to have a clear idea of the separations you want the column to perform, and don't hesitate to ask the vendor to confirm the column is suitable , perhaps even by running samples or providing an equivalent analytical column.
I've found it's alway a good strategy to negotiate the price of the preparative column, and then ask the vendor to provide guard/analytical columns at extremely-discounted rates.
The more you can fractionate unwanted junk out of your extracts before loading, the more good stuff you will be able to load each time. Watch the pressure changes during loading/running in analytical-scale runs, as repeat preparative runs can magnify increases.
Unless your extracts are really clean, I would strongly recommend purchasing a guard column with significant capacity to retain any really-sticky gunk from your natural product extracts. Guards are always cheaper to replace than columns of the same packing.
Depending on your natural products and extract, it may be preferable to perform a preliminary HPLC run to crudely fractionate the extract, and then a second run to resolve and purify closely-eluting compounds.
You can get good indications of sample cleanup and mobile phases from literature reviews using similar natural products.
Please keep having fun,
Bruce Hamilton