I tend to ultrasonic degas all premixed solvent mixtures, whether containing buffer or ion-pair salts or not. So, even if I vacuum filter a mobile phase with dissolved salts, I will also ultrasonicate it.
The reason for this is, unlike Hans, I don't have a vacuum system that could hold 1 litre Duran/Schott bottles, and I only have one of those bottles that is vacuum rated ( much more expensive than standard ones ).
If I filter through the Millipore 47mm all-glass system, I still have to transfer the filtrate to the 1L mobile phase reservoir, so quickly putting it into an ultrasonic bath removes at least some freshly-dissolved gases.
I agree that helium sparging is the best, especially for systems without an in-line degasser and using the same mobile phase repeatedly, eg on ion-chromatographs.
The main reason I prefer ultrasonication is that there are no sinters or vacuum lines that could contaminate the mobile phase. I have a 100 watt ultrasonic bath that really shakes those gases out, and can hold two 1 liter Duran bottles.
The ultrasonic bath was fairly cheap ( about US$150 ), and I think they are mainly used for helping to sterilize the implements of torture used by dentists. I recommend a high power unit, but be prepared for it to shatter autosampler vials if you also use it to help sample dissolution.
The "holes in aluminium foil" test gives a reasonable indication of the energy available.
With regard to single solvent mobile phases, I don't bother too much, unless I believe they will be an issue - such as with methanol-water gradients, and some normal phase mixtures. I rely on the instrument in-line degasser for those. But I should note that I have my Agilent 1100 degasser set for continuous operation, rather than the default mode.
The mobile phases I find giving the most pump problems are normal phase mixtures and aqueous-solvent ion-pair mixtures, which are isocratic.
The normal phase mixtures tend to give me the most problems with bubbles forming in some older UV detectors, but modern systems seem to have eliminated most detector bubbles..
With regard to filtering HPLC solvents, I assume they are OK from most reputable manufacturers, but may perform some confirmatory tests on new suppliers.
The easiest way to check for really bad contamination is to accurately preweigh two filters, stack them together, filter the solvent through them, allow them to air dry and reweigh. You should examine the upper surface of both using a binocular microscope. The lower filter also enables you to easily correct for weighing/drying variations without getting into a really fussy methodology....
Please keep having fun,
Bruce Hamilton