So you need to make the protocol more specific for your analytes, which hopefully are retained by both hydrophobic interaction and ion exchange. You load your analyte standards, on some 10 cartridges under acidic conditions. Next you wash with a basic solution (ammonia) containing increasing amounts of methanol: 10%, 20%, 30%... and measure the recovery of the analytes. You should get an S-shaped profile. Now you know that you can make additional washes at alkaline pH until the point of the methanol concentration just before the breakthrough of your analytes, and you can elute at alkaline pH with a methanol concentration just above the concentration where all analytes elute.
Now you improve your protocol as follows:
Load at acidic pH.
Wash under acidic conditions with methanol.
Reduce the methanol concentration to just below the breakthrough point.
Elute polar basic impurities now with this methnaol concentration at alkaline pH.
Elute your analytes at alkaline pH with a methanol concentration that is just beyond the point where you achieved 100% recovery.
Leave the rest of the basic interferences on the cartridge.
More work, but the results are likely to be better.
Also, we may need to look at improving your LC protocol, if this does not give you much improvements. However, I am reasonably confident that you will get improved results. The protocol has been proven to work on Oasis cartridges.