as the chart in the link shows that Waters ZQ2000 signal/peak area was 1) drifting and 2) had some abnormal variation when the same solution was injected multiple times. One vial was only injected once.

this was a gradient from 75%B to 100%B in 2.5 min. A: 0.1% acetic acid, B: ACN(no acid). 0.5mL/min, Water BEH 50x2.1 1.7um column, 35 C. Diluent: ACN. Analyte is a zwitterion.

I hope some one can tell me if this is normal.If not what I can do to improve it.

Thanks in advance
Excel
for system suitability testing on our LC-MS-MS (Waters 2790-ABI/4000) we used to use a <5% acceptance criteria on peak area ratio (API/IS).
Good judgment comes from bad experience, and a lot of that comes from bad judgment.
For an autosampler precision qualification test once (on an LC-MS-MS system), I put a UV detector between the LC and mass spec, injected caffeine, and monitored both responses. By UV, RSD ranged between 0.1 and 0.2%. By MS, RSD was 2 to 3.3%. Convinced me to use UV detectors to qualify autosamplers....which is not at all the same as setting system suitability parameters on an LC-MS-MS method.
All standard disclaimers apply. My posts are my opinions only and do not necessarily reflect the policies of my employer.
You also have to ask...

How much precision does your autosampler give you????

Alp
19 posts Page 2 of 2

Who is online

In total there are 35 users online :: 2 registered, 0 hidden and 33 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: Ahrefs [Bot], Google [Bot] and 33 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry