Do you have the option of temporarily running your MS from a higher-purity N2 cylinder, to see if that makes a difference?

any ideas how the N2 generator could be the source of contamination. Wouldn't you need a solvent to come into contact with N2 parts to leach the plasticizers out? The only liquid is water vapor from the air.

Tedinelkgrove,

I'm afraid I have no other N2 to switch to but the one from the N2 generator. I will try to mount the trap again in the system, though. My problems seem to have started somewhere around the time it was removed, so it might be a source of at least some problems.
Dejan Orcic
Asst. prof.
Department of Chemistry, Biochemistry and Environmental Protection
Faculty of Sciences, Novi Sad, Serbia
Tedinelkgrove,
What happened after you reinstalled the trap? Did the 214 disappear? I am having the same issue with my instrument, I have a LTQ Orbitrap XL. The 214 appeared about a year ago and we have tried everything we could think of for troubleshooting but the 214 is always present.
Celine
19 posts Page 2 of 2

Who is online

In total there are 32 users online :: 3 registered, 0 hidden and 29 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahrefs [Bot], Bing [Bot], Semrush [Bot] and 29 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry