1. Is this a USP method ( sounds like it with GF and IF ), if so which one?
Does the method specify or permit dilution with mobile phase. I would, if I could - refer 5 below.
2. What sample volume are you injecting?. Are you ensuring that it is cooled before dilution and filtration.
3. What happens when the system is just running with no injection, is the pressure is increasing - you suggest that it is, and it isn't. I would confirm that the pressure is stable with injections of mobile phase. If it isn't, then you mobile phase requires better filtration.
4. What happens to the pressure when you inject appropriately diluted, and cooled - if necessary, samples of each of the components of the sample solution ( Gastric Fluid, Intestinal Fluid, Buffer ), and then the unused blend used for sample solutions. My suspicion is that they should be OK in 70% aqueous, but you should check.
5. It seems that your samples have something that is precipitating in the mobile phase, so the next step may be to dilute your samples with mobile phase, allow to stand for a few minutes ( 5 - 10 ) before filtering. Then ascertain if the pressure still increases on injection.
6. If you have not got aguard column, I would be useful to add one, just to protect the column for dissolution media junk.
Please keep having fun,
Bruce Hamilton.