One of the big problems is the lifetime of a VOCARB trap in an OI. I don't know the current status of this issue, but a few years ago OI manufactured all of the traps for its instrments other than the VOCARB, which I believe is a #11 trap. Supelco had the VOCARB as a proprietary packing, and no other companies were allowed to manufacture the VOCARB trap. That isn't a problem with the Tekmar since it uses an external thermocouple to measure the trap temperature, but with the OI it is a big problem if the thermocouple is not welded to the trap correctly. IF the weld is not done properly the temperature measured is wrong, which can have a huge effect on trap lifetime and performance. When you inspect an OI trap the thermocouple wires should form a V which just touches the trap. If the wires cross above the weld, or if the two wires are actually welded separately the trap is actually much hotter than the measured temperature. The worst case I have heard of is one where the trap lasted two days, and was black on the outside when removed. Not all of the Supelco traps are bad, but it's hard to tell until you actually use one. I always have used an OI #10 trap for GCMS work.
The need for desorb preheat is related to trap heating rates. If the trap is heated rapidly there is no need for trap preheat. The OI has never really needed trap preheat, and it is much less important for newer Tekmars than for older ones.