Posted: Thu Jun 11, 2009 8:54 pm
Obviously, if required, prepare according to compendial instructions which often don't involve pH meters and take me about 30 - 60 minutes to set up, but otherwise for research work I perform the following.Please just state exactly how you prepared your buffers, so that it is known whether you are lazy or not.
Bruce, you prepare the dil. acid/base everytime also, or are they stable in your hands? Maybe you are not so lazy after all?
On a more serious vein: Some of the more stable modern pH meters can be less accurate than the older ones, where one has to have a bit of patience.
Also, I just wonder how you people do this adjusting the pH with an acid/base solution. When I did that I almost always overshot, then needed more buffer sol. then overshot again . . . . Do you hang the electrode into your solution?
I guess I am not only lazy, but inept to boot.
I'm not suggesting that this is desirable or best parctice, but it's quick and expedient...
1. Transfer electrode from storage solution, check calibration, rinse with Milli-Q water, and store in 400 ml of Milli-Q water for at least 30 minutes before use.
2. Weigh salt into 30 ml plastic bottle on 4 figure balance.
3. Add to calibrated 500ml Duran bottle, rinse plastic bottle with Milli-Q
4. Make up to about 98% volume with Milli-Q.
5. Ultrasonicate to dissolve and degas.
6. Add magnetic follower and electrode ( on holder arm ) and start stirrer.
7. Using a 1 ml pipettor ( less preferred is pasteur pipette ), immediately add 10% acid or base solution to desired pH - takes about a minute. I can't see how you would overshoot, as most buffers require much more of the acid or base as they approach the desired pH.
8. Remove electrode ( in buffer for 1 - 2 minutes maximum ), rinse with Milli-Q and return to storage solution.
9. Make up solution to 500 mL mark - decant to container.
Typical time to prepare ( excluding a couple of minutes for initial electrode rinse at start of day ), from start to finish, is about 10 minutes.
Dilute acid ( H3PO4 or HOAc ) and base ( KOH, NH4OH ) solutions are stable for at least 3-6 months here.
The pH meter stays in calibration ( three buffer pH=4, 7, 10, with 2 used for low pH ) for months without adjustment. I've actually found modern electrodes to be faster responding, and more consistent, than older electrodes, but these days I'm seldom measuring biological samples.
Yes I'm very aware of the comments about not putting electrodes into buffers - hence I keep contact to minimum, but I found making up to a specific volume ( eg 480 ml ), then taking 10 mls aliquots for pH reading and dumping them, recording accurately the added volume of acid or base, and then correcting the final makeup volume rather tedious ( but easy to calculate ) and unnecessary.
Obviously adding the electrode to the buffer is undesirable, but makes for a quick preparation. I've not seen any difference when I've analysed samples using mobile phases that were prepared without contact with the electrode and as prepared above.
Please keep having fun,
Bruce Hamilton