Posted: Fri Jun 05, 2009 2:46 pm
If the two chroms were run with the same moph and really on the same flow rate the area of the analyte should be the same, with and without the column, unless you are coinjecting air or the solvent deviates from mobile phase (or other system interference is present without the column in place). Also the peaks with the column seem to be at to, so also might be system peaks.
What I am trying to say is that such experiments are seldomly taken care of by making single injections, one has to play around a bit, for instance with the length and inner diameter of the capillary which is present when the column is removed.
Nevertheless one can already see that something is crossly wrong here judging from the baseline of the "with column" run. .....Didn´t Uwe suggest that you may have the problem of too much, not too little retention?
One more hint: The peak obtained with the column, if it is at a rt where system peaks are unlikely to hide under it, can not have a higher area than the "without column" peak.
What I am trying to say is that such experiments are seldomly taken care of by making single injections, one has to play around a bit, for instance with the length and inner diameter of the capillary which is present when the column is removed.
Nevertheless one can already see that something is crossly wrong here judging from the baseline of the "with column" run. .....Didn´t Uwe suggest that you may have the problem of too much, not too little retention?
One more hint: The peak obtained with the column, if it is at a rt where system peaks are unlikely to hide under it, can not have a higher area than the "without column" peak.