- Posts: 2846
- Joined: Mon Aug 30, 2004 7:17 am
What I am trying to say is that such experiments are seldomly taken care of by making single injections, one has to play around a bit, for instance with the length and inner diameter of the capillary which is present when the column is removed.
Nevertheless one can already see that something is crossly wrong here judging from the baseline of the "with column" run. .....Didn´t Uwe suggest that you may have the problem of too much, not too little retention?
One more hint: The peak obtained with the column, if it is at a rt where system peaks are unlikely to hide under it, can not have a higher area than the "without column" peak.