AICMM, thanks for all informative posts. Not until now had I known the rotor matching issue. We got the valve and heating box from agilent, never reminded the potential issue.GCLeo,
By changing the loop you have reduced the pressure drop on the switch to inject. I would agree with Gasman that the better way is a 1mL loop with a 1:1 or 1:2 split. As an alternative, if you have an aux pressure controller available you can try a high volume injection using the Aux EPC. Rotors are all polymeric, by the way, with the material varying from PTFE to PEEK so you should try to match the rotor to the analytes of interest (and the pressure and the temperature....) Valco makes Agilent's valve bodies but Agilent has their own actuators in order to fit under the hood of the GC.
Now for the interesting question, regarding Peter's note on back pressure regulators. I have heard from good sources that you will have to matrix match standard to sample if you use back pressure regulation. I have not tried this myself so I am interested in comments. I have tried back pressure regulation on the sample line and you do improve the detection limit but you also smear out the peak. Other comments?
Best regards.
In fact, I'm now using the 1ml loop with 1:1 or 2:1 split injection. Eventually, I am curious how you guys avoid tailing peaks when you do large volume injection. Is it "starting temperature at lower-than-boiling point to condense the sample, and then a temperature ramp" thing? I didn't do so as I set it to get an inline analysis every two minutes.