Uwe, it's not just about noise, it's about ionization. The increased solvent flow suppresses ionization in ESI. I see reduced peak heights at flow rates beyond about 300 µL/minute.Ok, I reread your info. A delay volume of 1 mL is not bad at all, and we used to run very fast LC/MS on systems like this. You objected, based on the idea that your loose a lot of sensitivity when you do that. Unless you have a MS system from the stone age, this should not be the case. So let me ask you the question why you think that your sensitivity goes down when you run at 1 mL/min. There could be problem in the thought, rather than a real problem.
Peak area decreases 5-fold when you change the flow rate 5-fold, because the peaks elute 5 times faster. But this is not a loss in sensitivity, unless the detector noise increases drastically with flow. With the MS systems that I know, the noise goes up a little bit with increased flow, bot not 5-fold for a flow rate change from 0.2 to 1.0 mL/min. Only maybe 50% or so. So to run at 1 mL/min is not a real problem.
If ESI behaved "ideally" and did not have this suppression property, then LC peak area would be identical regardless of flow rate; at higher flow rates peak width would contract and height would increase. But this is not the reality - with increased flow rate on ESI, I see reduced peak width *and* reduced peak height.