Hi All,
2 weeks ago I took out SPME liner and this time left it in a glass bottle with the liner not in contact with plastic top.
And I have installed it again with a new septa and bingo...
Flat baseline - no DEP - with 1st run.
I will soak some of the plastic tube the liner came in and run the solvent when I start doing some liquid injections.
Thanks
WK

Posted: Thu May 28, 2009 1:32 pm
by WK
Hi All,
I have just re-inserted the liquid injection liner which had also been stored in a glass bottle with the original silanised wool and bingo too - a very good baseline straight away with a new septa!!!
Thanks
WK

Posted: Thu May 28, 2009 4:46 pm
by zokitano
WK,

I am glad that you finally found the solution for your contamination problem.
It really helps us to be more aware about the possible sources of contamination in our regular lab work

Thanks,
Zoran

Posted: Fri May 29, 2009 12:13 pm
by Peter Apps
Hi WK

Thaks for the feedback - useful to know.

Peter

Posted: Fri May 29, 2009 1:08 pm
by WK
Hi Peter / Zokitano,
I have a routine now.
I should store the liners back in the plastic sleeve to see if problems might re-appear but i don't have time to risk that yet!
WK

Posted: Mon Jun 01, 2009 8:21 am
by Peter Apps
Hi WK

Just another thought - you should probably communicate this to the manufacturers of the liners so that they can change their packaging.