You aren't giving a lot of informations about your analysis issues, but i'll try to help you anyway.
I guess you are doing simulated distillation analysis of crude oil in a GC/FID, probably with a metal column or an high temperature one (like inferno of phenomenex).
You are asking if you can inject crude oil directly instead of a CS2 diluted solution.
I think you are asking this because you find difficult to quantitate the low boiling fraction, the one eluting with CS2?
I'd stick with CS2 solution injection, because the column doesn't have the concentration capacity for that type of injection and you would be overloading the stationary phase with sample.
If your problem is the quantitation of the fraction eluting with CS2, just run a solvent blank before the sample and subtract it from the sample chromatogram, you should be able to quantitate low boiling fraction.
Davide
Thanks Davide,
we use CS2 as dilution as per software (chrome perfect) calculation, my question is, can I use sample directly without diluting it. if I do, then what would be the effect on wt % recovery
Regards
Akbar