method validation for different matrixes

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

2 posts Page 1 of 1
Hello,

I'm quantifying one active from three different matrix compositions as follows:

Matrix 1: Contains components A and B and the active
Matrix 2: Contains components A, B and C and the active
Matrix 3: Contains components A, B, C and D and the active

am I right saying that I can perform specificity, linearity, accuracy etc. only on matrix 3 and the method will be valid also for matrix 1 and 2?

Another thing; I'll be using 100 ul injection routinely. However rarely I may need to use larger injection volume e.g. for diluted samples. Which validation tests should I repeat for the high volume extension?

thanks

Precision should be carried out for each formulation.
Linearity and recovery can be combined.
Specificty should be handled very carefully. Degradation studies should be performed for individual formulation.
Solution stability-For each formulation.
Robustness can be performed in formulation containing all components.
2 posts Page 1 of 1

Who is online

In total there are 66 users online :: 0 registered, 0 hidden and 66 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 66 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry