What are peoples procedures when setting up and washing down HPLC Systems i.e. what solvents should be used for reverse phase columns and for how long? What should the system be flushed with before runs and after runs and for how long?
-
- Posts: 1890
- Joined: Fri Aug 08, 2008 11:54 am
Personally I flush C18 and other "standard" columns with 70% MeOH or 70% ACN (no additives, no formic acid etc.) for 10 column volumes (or often just 20min on a standard "stop" method). If a manufacturer has specific instructions for their column, it's good to follow them, but when the ACN crisis was happening I checked with a major manufacturer that MeOH is fine as a storage solvent in silica-based C18 even though ACN was the recommendation. I don't use 100% in case there is any tiny residual amount of anything salty and insoluble in 100% organic, that might precipitate.
Most of our systems get sufficient use that we never have to put them into "storage" conditions, but would expect to replace aqueous solvent every couple of days (so no time for bugs to grow). Where we do leave things in storage, it's usually 20% MeOH or similar; enough organic to prevent bugs, but otherwise aqueous and no corrosive additives. Again, we want to avoid long-term precipitation of anything salty (e.g. in inlet filters, which are surprisingly difficult to rinse totally free of salt).
Most of our systems get sufficient use that we never have to put them into "storage" conditions, but would expect to replace aqueous solvent every couple of days (so no time for bugs to grow). Where we do leave things in storage, it's usually 20% MeOH or similar; enough organic to prevent bugs, but otherwise aqueous and no corrosive additives. Again, we want to avoid long-term precipitation of anything salty (e.g. in inlet filters, which are surprisingly difficult to rinse totally free of salt).
2 posts
Page 1 of 1
Who is online
In total there are 103 users online :: 2 registered, 0 hidden and 101 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Bing [Bot], Semrush [Bot] and 101 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Bing [Bot], Semrush [Bot] and 101 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science