Inlet Conditions : Splitless, Temp 250°C, Purge Time 1.00, purge Flow 50.0, saver Flow 20.0, Saver Time 5.00 - Drilled Uniliner Direct injection (hole at top)
Injection 1µl manual
Column ZB- Wax Plus (30m x 0.32 mm ID x 0.25µm thickness)
Oven 80 (2mins) to 160°C @ 10°C/min, hold for 2.00min, then 50°C/min to 250 for 2 minutes
FID Detector 250°C (H2 25, Air 350, Make Up N2 29.0)
Carrier Gas Nitrogen @ 3.0ml/min
The MEG peak is very broad (but symetrical), whilst my DEG/Butan1,4 Diol (IS)peaks are OK but not fully resolved. (see attached chromatogram) Using a 50/50 mix of IPA and water as the matrix did not make any difference to the peak shape quality.
Having done a calibration the DEG has a correlation factor of 0.9996 of which I'm more than happy with, but the MEG is just 0.97. Will I just have to live with that or can anyone suggest anything to improve it?
Any suggestions??