I'm having an issue with my GC. It's an Agilent 6890 FID. I am analyzing for ethanol and methanol in water samples. I normally run a set of standards at the beginning and at the end of each run. Usually, the end standards are just a little low from sitting on the sampler after they have been pierced. No problem. Recently, I have been getting low recoveries (90%) on the beginning standards and close to 100% on the end standards. Why are my initial standards so low? I change the liner, septum and o-ring before every run. After the issue began, I clipped the guard column (no improvement) then I changed the gold seal-still no improvement.
Does anyone have any ideas?
Thanks for your help.