The second step of the method system performance check is to measure detector response to the various FAMEs measured relative to the 23:0 methyl ester internal standard. To put this in the language of the method, the instrument's "Empirical Response Factor" (ECF) for each analyte must be measured using a specific commercially-prepared standard, and the deviation from the "Theoretical Response Factor" (TCF) calculated. ECFs must not deviate more than 3% from the TCFs for saturated and monounsaturated FAMEs.
The ECFs I measure on unsaturated and monounsaturated FAMEs are between 7 and 9% lower than the TCFs, so my system does not pass. If this criterion is not met, the method says "it is suggested the GC parameters (e.g., detector gases, carrier gas flows, and column choice, etc.) be reoptimized to achieve a difference of 3% or less before continuing with analysis." This is pretty vague! My question is, what factors are likely to affect the ECFs for these analytes? Where should I start "reoptimizing"?
Instrumental details:
30m x 0.25mm x 0.25 um FAMEWAX (Restek #12497)
Split/splitless inlet at 235C
1 uL injected by autosampler, split ratio 1:100
Helium carrier at constant flow of 2.4 mL/min
Oven program: initial 170C, 1 C/min ramp to 225C
Detector temperature 325C
Detector gas flows are not specified in the method; I'm using 30 mL/min hydrogen, 400 mL/min air, and helium as makeup (total of carrier + makeup 25 mL/min).
Thanks!