- Posts: 6
- Joined: Sun Mar 20, 2011 10:33 am
Advertisement
174 95 ratio in BFB tune fail
Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.
3 posts
Page 1 of 1
Hi, I am performing BFB tuning for my TD-GCMS. I keep getting above 100% for my 174/95 ratio. Is there a way to lower it to between 50-100%? I read somewhere that by increasing the ion energy, the ratio could pass? Is "ion energy" the "EIEnergy" in the agilent manual tune window? If so, how much should I increase??
-
- Posts: 880
- Joined: Fri Jan 04, 2008 3:03 pm
Yes, you can ramp the emission current and pick the max. Also try to get 131 and 219 to between 40- 50% of 69 in your manual tune.
-
- Posts: 6
- Joined: Sun Mar 20, 2011 10:33 am
Hi Big bear, thanks for the help! Much appreciated!!
3 posts
Page 1 of 1
Who is online
In total there are 39 users online :: 1 registered, 0 hidden and 38 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm
Users browsing this forum: Google Adsense [Bot] and 38 guests
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm
Users browsing this forum: Google Adsense [Bot] and 38 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science