issue : Retention time shifting in ion chromatography

Discussions about IC and related topics

10 posts Page 1 of 1
issue : Retention time shifting

in ion chromatography

instrument: metrohm

column : metrosep A supp1 250

mobile phase : 0.318 gm in 1000mL water

detecion: sulfate

initial Rt :24(very 1st injection) and after 100 injections its 21

can any one suggest any solution for it

thanks in advance
---------------------------
With Warm Regards
Pr. D.Sudhir Kumar M.Pharm
Research Associate-IV
AUROBINDO PHARMA
---------------------------
is the RT 24 throughout the first 100 injections and then suddenly 21 min at 100? Or is it a gradual change from 24 min to 21 throughout 100 samples? Does it go back to 24 on its own or do you have to do any maintenance (e.g. backflush column etc.).
~Ty~
is the RT 24 throughout the first 100 injections and then suddenly 21 min at 100? Or is it a gradual change from 24 min to 21 throughout 100 samples? Does it go back to 24 on its own or do you have to do any maintenance (e.g. backflush column etc.).
Through out the analysis/or till the 100 inj(its not the exact figure) runtime was 24 and for the next time when i performed the analysis its 21 I tried with diff systems, diff M.phase, at-last when i tried with new column it was again 24
and one more thing is that the old column which i was getting 21 reduced its Rt.
---------------------------
With Warm Regards
Pr. D.Sudhir Kumar M.Pharm
Research Associate-IV
AUROBINDO PHARMA
---------------------------
How long did the column sit between those sets of runs? what was the storage buffer (and what did the manufacturer recommend). It looks like your column is dying during storage.
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
How long did the column sit between those sets of runs? what was the storage buffer (and what did the manufacturer recommend). It looks like your column is dying during storage.
the mobile phase which i run is sodium carbonate anhydrous(0.3 gms in 1000mL water) buffer, and its the storage solution recommended by the column vendor
---------------------------
With Warm Regards
Pr. D.Sudhir Kumar M.Pharm
Research Associate-IV
AUROBINDO PHARMA
---------------------------
All the symptoms point to the column deteriorating, so the next step would be to contact the vendor.
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
what is the sample matrix?
back pressure on the column (initially and with reduced RT?
Dr. Markus Laeubli
Manager Marketing Support IC
(retired)
Metrohm AG
9101 Herisau
Switzerland
what is the sample matrix?
back pressure on the column (initially and with reduced RT?
yes the pressure is gradually increasing
initial - 7.8
present - 9.4
---------------------------
With Warm Regards
Pr. D.Sudhir Kumar M.Pharm
Research Associate-IV
AUROBINDO PHARMA
---------------------------
what is the sample matrix?
back pressure on the column (initially and with reduced RT?
the sample matrix is enoxaparin injection
---------------------------
With Warm Regards
Pr. D.Sudhir Kumar M.Pharm
Research Associate-IV
AUROBINDO PHARMA
---------------------------
Are you using a Guard column?

If yes, how is the back pressure without Guard?
Did you replace the guard column?

If no, check with Metrohm India about column regeneration on this application.
Dr. Markus Laeubli
Manager Marketing Support IC
(retired)
Metrohm AG
9101 Herisau
Switzerland
10 posts Page 1 of 1

Who is online

In total there is 1 user online :: 0 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 1 guest

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry