FID broken on Agilent 7890 GC
Discussions about GC and other "gas phase" separation techniques.
5 posts
Page 1 of 1
Whenever I put on my Agilent 7890A GC it always display "FRONT DETECTOR (FID) BROKEN". What could be responsible for this
-
- Posts: 71
- Joined: Thu Nov 04, 2010 12:14 pm
Loose/disconnected cables.
Faulty electronic boards.
Faulty electronic boards.
-
- Posts: 3
- Joined: Tue Jun 19, 2012 11:54 am
Which of the boards could this be so that I dont just grope in the dark
-
- Posts: 3210
- Joined: Thu Sep 02, 2004 7:28 pm
main mother board
FID electrometer board
best wishes,
Rod
FID electrometer board
best wishes,
Rod
-
- Posts: 373
- Joined: Mon Sep 13, 2010 8:28 am
Shortened heater.
Disconnected thermocouple.
Shortened to chassis collector electrode or dirt on collector.
Disconnected thermocouple.
Shortened to chassis collector electrode or dirt on collector.
5 posts
Page 1 of 1
Who is online
In total there are 38 users online :: 1 registered, 0 hidden and 37 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm
Users browsing this forum: Amazon [Bot] and 37 guests
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm
Users browsing this forum: Amazon [Bot] and 37 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science