When I have time I will try to look up that information. I believe the retention time is very close to Kr.
Edited ON: the retention of methane on a C-1000 column is about 4 times the retention of nitrogen.
VERY CLOSE to Argon. Edited OFF
I know you need BOTH columns to get the separation you need.
The most simple answer is to determine how much time the Kr is retained behind nitrogen on a 2 or 3ft column and then use a MSieve 5A column that will retain methane half that time behind nitrogen.
Since the C-1000 is stronger retaining Kr behind nitrogen than a MS5A is retaining methane behind nitrogen, you need a short C-1000 column and a long MS5A column.
Quickly offering a
guess ,
I think a 8 ft MS 5A will require a 3 ft Carboxen 1000.
But I would buy a 1, 2 and 3ft Carboxen 1000 (giving you 1, 2, 3, 4, 5, and 6ft combinations);
and to go with your present 8ft column
a 4ft and a 10ft MS 5A (giving you 4, 8, 10, 12, 14, and 18 ft. combinations).
[ 5 packed columns at a price that perhaps a single 60 meter capillary would cost you ]
And don't forget the 1/8 inch bored through SS unions to connect the columns together. One of these combinations should do the trick, no?
best wishes,
Rod