TCE and PCE screening method
Discussions about GC and other "gas phase" separation techniques.
3 posts
Page 1 of 1
Hello, I am trying to set up a quick screening analysis to determine if a sample has TCE or PCE in it. I am using an ECD with a Zebron Inferno column (30 m). I have tried using different solvents to determine where the TCE elutes but I can't seem to find it. I have an FID as the front detector and the ECD is the back detector and I am using a y connector to split the Inferno column just before it reaches the detectors. The FID is being used to screen for diesel and oil in samples. Any help would be greatly appreciated!
-
- Posts: 3210
- Joined: Thu Sep 02, 2004 7:28 pm
While you did not tell us [please do be more specific if I do not guess correctly]
I ASSUME
you are using a biodiesel type column (like a 0.1um film and 0.25mm ID column)
Were all the peaks of the solvents you tried coming out very early in the SAME place?
If the answer is YES. Then guess where TCE elutes !
If I guessed incorrectly, well blame yourself for giving so little information for the Forum to work with.
best wishes,
Rod
I ASSUME
you are using a biodiesel type column (like a 0.1um film and 0.25mm ID column)
Were all the peaks of the solvents you tried coming out very early in the SAME place?
If the answer is YES. Then guess where TCE elutes !
If I guessed incorrectly, well blame yourself for giving so little information for the Forum to work with.
best wishes,
Rod
-
- Posts: 1773
- Joined: Fri Feb 09, 2007 10:33 pm
pacerlaser,
A suggestion. It sounds like you have PCE and TCE on hand, so why not put a uL of each into a beaker and let them evaporate and then shoot the headspace. The other solvent you might think about is DMSO which elutes rather late (relative to lots of other solvents.
Best regards,
aicmm
A suggestion. It sounds like you have PCE and TCE on hand, so why not put a uL of each into a beaker and let them evaporate and then shoot the headspace. The other solvent you might think about is DMSO which elutes rather late (relative to lots of other solvents.
Best regards,
aicmm
3 posts
Page 1 of 1
Who is online
In total there are 47 users online :: 1 registered, 0 hidden and 46 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm
Users browsing this forum: Baidu [Spider] and 46 guests
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm
Users browsing this forum: Baidu [Spider] and 46 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science