Any ideas as to what could cause a static headspace injection to produce peak areas roughly 60-80% higher than the other injections in the series?
For example, in a 6-injection series of a std. solution containing 12 residual solvents, 5 out of the 6 injections will be right on top of each other with %RSDs <5 for all analytes. The odd injection will have peak areas up to 80% higher than the other five.
Replacing the syringe seemed to cure the issue for now, but I can't understand how a leaky syringe could cause a high injection but a leak inevitable causes low values - and an inconsistent leak can cause low values in the majority of runs, leaving the "true" leak-free values as apparent high outliers .
We're using a CombiPal autosampler and Agilent 7890.
If this is the same problem as you posted in your other thread then all the suggestions there also still apply.
Look for patterns in when the high values occur - first of a batch ?, last of a batch ? etc
Peter