I currently use an Agilent HP-ULTRA 2 to quantify Polyethylene Glycol in Carbowax.

I'm working on a method that will look at quantifying Propylene Glycol in a HydroGel. I've been getting good results from the PoraPLOT-Q column.

To avoid disturbing the system routinely, I was wondering if there is a suitable column I could use for both tests?

Regards.
IF
you can use the Agilent HP-ULTRA 2 for the propylene glycol
then you have the single column you desire.

I doubt there is any other possibility.

best wishes,

Rod
IF
you can use the Agilent HP-ULTRA 2 for the propylene glycol
then you have the single column you desire.

I doubt there is any other possibility.

best wishes,

Rod
Initial trials showed that the HP-ULTRA 2 didn't work with the PEG.

Based on that and what you say; it looks like we have no choice.

Thank you.
3 posts Page 1 of 1

Who is online

In total there are 32 users online :: 1 registered, 0 hidden and 31 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: Google [Bot] and 31 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry