I have an ion chromatograph for separating amino acids using Na exchange. The system is very close to validation except for one little detail. I can't get the peaks to stay at a consistent retention time and they sway back and forth within their window. There is variability with cysteic acid, found at the void volume. So my problem is, I have shifting RT at the VV.

The most obvious cause is a leak somewhere. Ladies and Gents, I swear to you I have looked this system up and down and cannot find a leak anywhere.

My best guess is bubbles in the eluents. There is no degassing system, although the manufacturer has sent me an upgrade that I will have to install myself. Right now the eluent are in inverted IV bottles that I've had to puncture with a pinhole just to equalize pressure (a design problem). I think air is dissolving into the eluents where later they fiendishly wreak havoc on my system.

If anyone has any other suggestions of where I can look, much appreciated. If anyone has questions I'll be happy to clarify as much as I can safely elaborate.

Thanks in advance

I had this kind of problem many years ago. It turned out that the proportioning valve had a slight leak and the pH of the eluent was dancing around. It only affected the first couple of peaks. This was a brand new HPLC too! I have also seen even minor leakage in a checkvalve cause wandering retention times, but it affects all peaks, not just the first one. These kind of leaks are internal and don't drip where you can see them.

I doubt that you have a bubble. This would cause erratic pressures, bad baselines and affect all peaks. Even so, the degasser upgrade is a good idea.

By the way, cysteic acid is not the void volume marker. It is actually separated by ion exclusion. Void volume = excluded volume + pore volume. Cysteic acid elutes near the excluded volume; urea elutes near the void.
Mark Tracy
Senior Chemist
Dionex Corp.

Nuts. Not exactly the best news I could have hoped for.

Thanks for pointing out an erratic check valve. This might be the case, but often when I reprime each eluent I get some very small bubbles coming from the valve, so I think I'll look at this first.
By the way, cysteic acid is not the void volume marker. It is actually separated by ion exclusion. Void volume = excluded volume + pore volume. Cysteic acid elutes near the excluded volume; urea elutes near the void.
I did not know that! Thank you for that tip. I guess what you're saying is that cysteic acid actually elutes before the void. That would make sense; my baseline dip follows my component and I just assumed that CYS-Ac was at the head of the slug.
3 posts Page 1 of 1

Who is online

In total there are 18 users online :: 0 registered, 0 hidden and 18 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 18 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry