Lightnig the FID on Shimadzu GC-17A

Discussions about GC and other "gas phase" separation techniques.

4 posts Page 1 of 1
Q: Is there a fool-proof way to get this FID lit? Our procedure has been to have N2/50 kPa, H2/50 kPa, air/50 kPa when running and to hike the H2 up to 100 kPa for lighting. But this is really hit and miss, and today I am having no luck with that whatsoever.

Suggestions?

Bob Hanson
St. Olaf College
When igniting turn the make up gas off, and turn the air down to about half its running value. Check your flow rates in any case with a flow meter - the pressure readings do not tell you what you are actually putting through the detector.

Peter
Peter Apps
When was the last tiem the glowplug was replaced? Also check the filament is in contact with the gas stream
Where can I buy the kit they use in CSI?
Check the flame jet , it may be partially blocked.
Also check the ferrule at the detector side and tighten if necessary.

65 Kpa for Hydrogen and 50 Kpa for Air are standard pressures for optimum flow rates for Shimadzu.
4 posts Page 1 of 1

Who is online

In total there are 46 users online :: 0 registered, 0 hidden and 46 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: No registered users and 46 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry