Inconsistent areas for gas injections.

Discussions about GC and other "gas phase" separation techniques.

4 posts Page 1 of 1
The areas for gas injections to Agilent GC with SCD is giving very inconsitent areas. It varies from 10,000 to 90,000 with each run. When liquid sample is injected the area difference is mostly within 1000. The injections are done at 40:1 split ratio.
What other information will be needed to suggest what the problem is?

thank you
More information is needed.

Which Agilent model of GC.

Are you using a valve for gas injection

If you are using a valve, are you using pre-treated tubing and inlet such as Sulfinert, Silcosteel or Ultimetal treatments.

Gasman
We are using Agilent Gc, 7890A model.
Yes there is a gas valve used, and all are pre treated.
What are the analytes? What solven are you using when you make a liquid injection? At what temperature is the gas valve kept?
Where can I buy the kit they use in CSI?
4 posts Page 1 of 1

Who is online

In total there are 45 users online :: 0 registered, 0 hidden and 45 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: No registered users and 45 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry