Aqueous HCl as Mobile Phase Constituent
Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.
9 posts
Page 1 of 1
I'm using 4 mM aqueous HCl (pH = 2.5) as the primary (95%) mobile phase constituent for a polar C18 separation. Is this liable to corrode the pumping system? I'd be interested in anyone's personal experience (good or bad) using HCl in this manner. Thanks!
-
- Posts: 2916
- Joined: Mon Aug 30, 2004 10:19 pm
Chlorides in general and HCl in particular are supposed to attack stainless steel. I have seen people use HCl in so-called inert systems, but not in standard steel systems.
-
- Posts: 2846
- Joined: Mon Aug 30, 2004 7:17 am
We have been through this many times, extensively!
But: Take a piece of stainless (preferably one section shined) and dump it into your HCl. Let it stand for some weeks, occasionally observing what happens (and please tell us about the results). I know that at higher HCl conc., stainless turns black very fast ("normal" iron, a nut, turned red brown). At a concentration of 0.15M Cl- at neutral or higher pH didn´t do anything in months to the stainless used here. I have used PBS (0.15M phosphate, 0.15M NaCl, pH ~ 7.2) for years without any problems (don´t leave the stuff on there longer than needed, though).
But: Take a piece of stainless (preferably one section shined) and dump it into your HCl. Let it stand for some weeks, occasionally observing what happens (and please tell us about the results). I know that at higher HCl conc., stainless turns black very fast ("normal" iron, a nut, turned red brown). At a concentration of 0.15M Cl- at neutral or higher pH didn´t do anything in months to the stainless used here. I have used PBS (0.15M phosphate, 0.15M NaCl, pH ~ 7.2) for years without any problems (don´t leave the stuff on there longer than needed, though).
-
- Posts: 144
- Joined: Mon Aug 30, 2004 10:12 pm
Is there a reason that you are not using phosphoric acid which we all know to be safe to use in LC systems over extended periodes of time?
Bill Tindall
-
- Posts: 5
- Joined: Fri Jul 29, 2005 12:31 am
I've been thunking my forehead every time I think about it now. I'm too far down the road to just switch to phosphoric at this point...unless corrosion is a real issue. But you are right, I know of no good reason why phosphoric acid would not work in my application.Is there a reason that you are not using phosphoric acid which we all know to be safe to use in LC systems over extended periodes of time?
-
- Posts: 2916
- Joined: Mon Aug 30, 2004 10:19 pm
Hmmh, not a good spot to be in...
If the method will need to be run on other systems, I would follow the general advice given by the engineers who design such systems and rework the method. No question.
If the method will run only on my own system, and if it is a lot of work to redevelop and requalify the method, I might take the risk and run it on my machine. 4 mM? Maybe that is not enough to do any serious harm...
If the method will need to be run on other systems, I would follow the general advice given by the engineers who design such systems and rework the method. No question.
If the method will run only on my own system, and if it is a lot of work to redevelop and requalify the method, I might take the risk and run it on my machine. 4 mM? Maybe that is not enough to do any serious harm...
-
- Posts: 5
- Joined: Fri Jul 29, 2005 12:31 am
I left a shiny column fitting in 10 ml of the aqueous mobile phase per your suggestion. After 3 weeks no noticeable physical changes. I also analyzed the solution for iron and detected none (<1 ppm). I thank you and others for your helpful comments.We have been through this many times, extensively!
But: Take a piece of stainless (preferably one section shined) and dump it into your HCl. Let it stand for some weeks, occasionally observing what happens (and please tell us about the results). I know that at higher HCl conc., stainless turns black very fast ("normal" iron, a nut, turned red brown). At a concentration of 0.15M Cl- at neutral or higher pH didn´t do anything in months to the stainless used here. I have used PBS (0.15M phosphate, 0.15M NaCl, pH ~ 7.2) for years without any problems (don´t leave the stuff on there longer than needed, though).
-
- Posts: 310
- Joined: Mon Aug 30, 2004 4:56 pm
Ron,
As Hans says, we'd been through this discussion many times and I haven't chimed in as the discussion regarding compatibility of stainless steel with chloride containing eluents always seems to go nowhere. But given your post, I think it's necessary to add a couple of clarifying points. First off, the test you described is inadequate for determining what happens in the real instrument situation. Corrosion in metallic systems is generally caused by a combination of redox potentials between different metal components, loss of the passivation layer and contamination of the metal surface with tiny metal fragments derived from the machining process. The beaker experiment you did doesn't adequately test the real situation in the instrument. Do the following: swage a ferrule onto a stainless steel tube and then store it in HCl. In most cases, you'll see corrosion occurring, especially at contact points where the swaging component intersects the surface of the ferrule even though the individual components won't necessarily show any sign of corrosion if tested alone. Secondly, the passivated layer on the surface of stainless steel is not completely inert to cycles of exposure to acidic and basic solutions. If you alternate exposure of the stainless steel component to alternating extremes of pH, you will also generally observe corrosion with corrosive eluents. For this reason, it's generally inadvisable to use corrosive reagents such as HCl without taking extreme care to avoid corrosion problems. If you really find it necessary to use stainless steel in conjunction with HCl, it's highly advisable that you frequently passivate your instrument, avoid eluents containing chelating agents and high pH eluents. A better choice, though, is to use a noncorrosive acid such as methanesulfonic acid.
As Hans says, we'd been through this discussion many times and I haven't chimed in as the discussion regarding compatibility of stainless steel with chloride containing eluents always seems to go nowhere. But given your post, I think it's necessary to add a couple of clarifying points. First off, the test you described is inadequate for determining what happens in the real instrument situation. Corrosion in metallic systems is generally caused by a combination of redox potentials between different metal components, loss of the passivation layer and contamination of the metal surface with tiny metal fragments derived from the machining process. The beaker experiment you did doesn't adequately test the real situation in the instrument. Do the following: swage a ferrule onto a stainless steel tube and then store it in HCl. In most cases, you'll see corrosion occurring, especially at contact points where the swaging component intersects the surface of the ferrule even though the individual components won't necessarily show any sign of corrosion if tested alone. Secondly, the passivated layer on the surface of stainless steel is not completely inert to cycles of exposure to acidic and basic solutions. If you alternate exposure of the stainless steel component to alternating extremes of pH, you will also generally observe corrosion with corrosive eluents. For this reason, it's generally inadvisable to use corrosive reagents such as HCl without taking extreme care to avoid corrosion problems. If you really find it necessary to use stainless steel in conjunction with HCl, it's highly advisable that you frequently passivate your instrument, avoid eluents containing chelating agents and high pH eluents. A better choice, though, is to use a noncorrosive acid such as methanesulfonic acid.
-
- Posts: 2846
- Joined: Mon Aug 30, 2004 7:17 am
Well, maybe I oversimplified tests a bit, what I use is a piece of the piping, from my system, polished in one place, the rest having flaws and rough cut edges. In short, I have never seen any corrosion, tarnishing, etc. anywhere, in connection with Cl- (max. 0.15M) at a pH of 7 or higher (years of contact). I don´t recall what acidic Cl- conc. I tested, but any of the HCl solutions I tried failed the test (turned black).
So I wonder, Ron, do you see any signs of oxidation or corrosion on the visible parts of your HPLC after having run your HCl?
Chris, I wonder whether the corrosion you mention was seen even with something like 4mM HCl. Actually an answer is not so important as I agree that it´s prudent to avoid acidic Cl-.
So I wonder, Ron, do you see any signs of oxidation or corrosion on the visible parts of your HPLC after having run your HCl?
Chris, I wonder whether the corrosion you mention was seen even with something like 4mM HCl. Actually an answer is not so important as I agree that it´s prudent to avoid acidic Cl-.
9 posts
Page 1 of 1
Who is online
In total there are 23 users online :: 1 registered, 0 hidden and 22 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Bing [Bot] and 22 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Bing [Bot] and 22 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science