Filament burn out at GCMS

Discussions about GC and other "gas phase" separation techniques.

6 posts Page 1 of 1
May I ask why the filament always burn out even no leakage when initial run? the burn out's filament is shown when the trap emission shown its abnormality after run several sample.

May I know the reason?

Thanks..:)
Well, maybe that is because of sample?

Do You switch off filament during the solvent elution? Maybe the analyte itself is reactive and present in high concentration.
Inject only low conc. samples and keep it OFF while solvent elutes. Filaments burn when they are ON with a lot of molecules to ionize.
Davide Balbo from Italy
How long should we set for the methanol's solvent delay?

Thanks for the fast reply.


Thanks.


Penny Tee
It depends on the column dimensions, GC oven temperature and gas flow.

In most cases 2.5-3 minutes is enough for 25-30 m columns.
Thanks.:)
6 posts Page 1 of 1

Who is online

In total there are 27 users online :: 2 registered, 0 hidden and 25 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: Google [Bot], Semrush [Bot] and 25 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry