Hi! I have set up a headspace SPME method on our GC-MS. I am quantitating a compd. using an internal standard. I have a calibration curve with r-square= 0.993. Now, for validation purposes I am running samples of known concentrations. But the values after quantitation are all over the place. What could be the cause of this? Where do I start to look for this problem. The samples are being prepared the exact same way and everything....
-
- Posts: 216
- Joined: Mon Apr 04, 2005 12:26 am
I don't understand your problem totally.
You say you have a good calibration curve. Is the matrix of your calibration curve the same as that of the real samples.
If so, doesn't make any sense. If the sample matrix is different than your matrix to prepare the standard curve, need to make your calibration matrix more like your standard matrix.
Maybe you are getting some matrix effect due to interferences. Might be hard to tell if just doing selected ion monitoring instead of a full mass scan. THe other possibility would be different recovery efficiencies of your analyte in your matrix versus your standard matrix.
You say you have a good calibration curve. Is the matrix of your calibration curve the same as that of the real samples.
If so, doesn't make any sense. If the sample matrix is different than your matrix to prepare the standard curve, need to make your calibration matrix more like your standard matrix.
Maybe you are getting some matrix effect due to interferences. Might be hard to tell if just doing selected ion monitoring instead of a full mass scan. THe other possibility would be different recovery efficiencies of your analyte in your matrix versus your standard matrix.
Sailor
-
- Posts: 20
- Joined: Fri Jul 08, 2005 9:30 pm
Thanks for your reply. I am running the samples for validation in the same exact matrix (red wine). Yet, I am not getting very good results when I quantitate the samples with amts against the standard curve. Doesnt make sense and I am wondering if there is something I am doing wrong as far as quantitating the samples goes. How should I go about quantitating these samples with the standard curve generated? I am new to all this!!
3 posts
Page 1 of 1
Who is online
In total there are 31 users online :: 1 registered, 0 hidden and 30 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Google [Bot] and 30 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Google [Bot] and 30 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science