Peaks often disappear due to contamination in the inlet so first clean your inlet and replace the liner. If that doesn't restore the peak(s), cut off one meter from the inlet end of the column. That is where sample crud condenses irreversibly changing the column activity.
Also, when it comes to sustaining chromatographic integrity regarding column contamination, what is the best method of recurring maintenance: solvent wash or baking?
The best practices for column maintenance are to change inlet liners often to minimize crud getting onto the column. Cutting on-meter increments from the front of the column is the second step. Flushing the column with solvents is the last resort and may or may not be effective.