Does anyone know what Instrument failure W2690/5@2 in the empower message window mean? it also says W2690/5 2 IEEE488 failure. What do I do to fix it?
empower failed to connect to w2690

- make sure IEEE488 is properly connected
- system is powerd on
- system is on line (>configure systems >?systems? >"status/online ok" (or left click on system, "bring online")

Tipp: if system is already listed online, maybe "take offline" then "bring online" again.
Hi,

Make sure the IEEE address for the alliance is per the software. Check the config of the alliance.
to expand on^ - there is a little battery in each component, when they die, the IEEE reverts to the default setting. If that happens not to match what you are looking for, or there's another component on the same LAC/E w/ the same address, nothing will work right.
Thanks,
DR
Image
4 posts Page 1 of 1

Who is online

In total there are 182 users online :: 1 registered, 0 hidden and 181 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: Google [Bot] and 181 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry