SAX for bromide

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

2 posts Page 1 of 1
Hi
I need to analyze bromide in water containing TriChloroBenzene residue. I use Parisil SAX (Whatman) column and KH2PO4 buffer as a mobile phase. Bromide is stronly interfered with the TCB residue. I am thinking to add an organic modifier to the mobile phase. Which column and mobile phase can you recommend?
That's a silica-based column, so solvent limitations are the same as those for any bonded-phase column. Adding either MeOH or ACN should make the TCB elute earlier. The other possibility is to change your buffer ionic strength. Increasing the salt concentration should make Br- elute earlier; decreasing it should make Br- elute later.
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
2 posts Page 1 of 1

Who is online

In total there are 177 users online :: 2 registered, 0 hidden and 175 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: Bing [Bot], Google [Bot] and 175 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry