Hi all,

I am trying to get Chemstation ver A10.. to communicate with my 5890 series II GC after upgrading the computer to Windows XP. We have a PCI GPIB card in the computer but only an HPIB card in the GC. Agilent thought this should work. In the configuration, there is only a option to put an address in for a GPIB card. The HPIB address doesn't seem to work. Anybody out there using this version of chemstation with the 5890? Do you need a GPIB card for the GC or did you get it to communicate with the HPIB.

Any help would be appreciated.

Thanks.

Carol

Run the VISA assistant and see what/if it detects your GC and what the IP address is: usually 15 by default for the GC and 8 for an autosampler. Check the GC config, should be in the same suite of software as the VISA, and make sure that your instrument IP here is the same as in the VISA.

Cards should probably work. Some versions of chemstation have OS compatibility issues though. For example, I still run a version of MS chemstation where the enviroquant is not compatible with XP/VISTA.

I'm assuming that you have either 82341 or 82350 GPIB board in the computer. You must set that up as in the instructions, install from the HPIO Libraries. Both get set up as if they are 82341 boards. You must configure manually, type in "hp82341" no matter which of the two boards you really have installed, and set the Bus Address to 30 (whatever that means). The Configuration Editor setting for an HPIB address was for older 82335 cards for Pentium 1 and older computers, don't use that.

The GC modules get configured same as in past, through Configuration Editor, usually with the "15" and "8" defaults described earlier.
Thanks guys for your assistance. No luck so far but will try out your suggestions and see what I get.

Carol

Re:

I'm assuming that you have either 82341 or 82350 GPIB board in the computer. You must set that up as in the instructions, install from the HPIO Libraries. Both get set up as if they are 82341 boards. You must configure manually, type in "hp82341" no matter which of the two boards you really have installed, and set the Bus Address to 30 (whatever that means). The Configuration Editor setting for an HPIB address was for older 82335 cards for Pentium 1 and older computers, don't use that.

The GC modules get configured same as in past, through Configuration Editor, usually with the "15" and "8" defaults described earlier.

This should work, let us know if you got it working
Hi raink91,

Yes, we did get it working. It actually turned out to be a problem with the cable in that it was not tight enough between the GC and computer. Once we got that straightened out, everything worked. No problems since.

Thanks for responding.

cmoral
6 posts Page 1 of 1

Who is online

In total there are 39 users online :: 0 registered, 0 hidden and 39 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: No registered users and 39 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry