negative peak in fid before methane
Discussions about GC and other "gas phase" separation techniques.
3 posts
Page 1 of 1
i am using an agilent gas chromatograph with tcd and fid.i am using helium as an carrier gas for the fid.i inject my gas directly in the fid and i am getting a negative peak before methane.this peak is not observed in the standard gas.even when i use a nucon gc, with nitrogen as the carrier i am facing the same problem.please help.thanks
-
- Posts: 3210
- Joined: Thu Sep 02, 2004 7:28 pm
what is different from your sample gas from your std gas?
Rod
Rod
-
- Posts: 2
- Joined: Tue Jan 03, 2012 2:58 pm
hi! my sample gas is the gas desorbed from subsurface soil samples using orthophosphoric acid.my standard gas does not have co2 in it.i injected another standard gas having co2 and i got the same signature as my sample.there was a dip before methane.i am using agilent 7890a with helium as the carrier in fid
3 posts
Page 1 of 1
Who is online
In total there are 40 users online :: 0 registered, 0 hidden and 40 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm
Users browsing this forum: No registered users and 40 guests
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm
Users browsing this forum: No registered users and 40 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science