I have been using the kromasil columns for a long time , I have been using the TEAP pH3.0 as the mobile phase,gradully i find that after a period of time shoulder peak will appear by using kromasil c-8 column ,yet this does not happend on the kromasil c-8,can anyone tell me
why is so ?
I have been using the kromasil columns for a long time , I have been using the TEAP pH3.0 as the mobile phase,gradully i find that after a period of time shoulder peak will appear by using kromasil c-8 column ,yet this does not happend on the kromasil c-8,can anyone tell me
why is so ?
As far as I know, kromasil c-8 should behave just like kromasil c-8. :wink:

If you really meant that the c-8 column is less stable than the c-18 column, then this is fairly typical. The shorter-chain bonded phase is more likely to be lost with low-pH cleavage of the silyl-ether linkage.
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
2 posts Page 1 of 1

Who is online

In total there are 27 users online :: 3 registered, 0 hidden and 24 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Google [Bot], kaladey84, Semrush [Bot] and 24 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry